Hi @anne_sa,
There are lot of options for MSA depending on the guideline you may refer to (AIAG, VDA, Bosch, ...).
What is common and critical for MSA is to differentiate part-to-part variation from measurement variation (equipment, operator, other external factors...) which can be calculated with %R&R in JMP (measurement quality), and also the accordance of your measurement variability depending on your specification (customer quality), which is calculated with P/T (Precision vs. Tolerance ratio, noted "Precision to Part variation" in JMP). The formulas are shown in my screenshots.
Both of these indicators should be <30% to have acceptable results (on the second link JMP provides some guidance about how to interpret the values), which means you want a batch variance at least 10 times higher than the rest of the variance (repeatability, operator and operator x batch).
You might find more infos here for JMP: https://www.jmp.com/support/help/en/16.2/index.shtml#page/jmp/about-the-gauge-rr-method.shtml#ww1925...
https://www.jmp.com/support/help/en/16.2/index.shtml#page/jmp/gauge-rr-option.shtml#
And more generally about assessing measurement quality : https://sixsigmastudyguide.com/measurement-systems-analysis/
This is also often described as "Type-II study", where you assess operator, equipment and measurement variation. But prior to this study, it is highly recommended to realize "Type-I study" where you keep one operator and one equipment, just to check measurement variability and that your measurement process is statistically in control and stable (assessment done with the calculation of Cp/Pp and Cpk/Ppk, and through the use of control chart).
I hope this first answer will help you,
Victor GUILLER
L'OrƩal Data & Analytics