cancel
Showing results for 
Show  only  | Search instead for 
Did you mean: 
  • JMP will suspend normal business operations for our Winter Holiday beginning on Wednesday, Dec. 24, 2025, at 5:00 p.m. ET (2:00 p.m. ET for JMP Accounts Receivable).
    Regular business hours will resume at 9:00 a.m. EST on Friday, Jan. 2, 2026.
  • We’re retiring the File Exchange at the end of this year. The JMP Marketplace is now your destination for add-ins and extensions.

Discussions

Solve problems, and share tips and tricks with other JMP users.
Choose Language Hide Translation Bar
Lien
Level II

Problem with graph

Hello dear JMP-Community!

 

I am trying to make a graph, which is pretty easily done in excel, in JMP. (see finished excel) 

Since I have only been working with JMP for about 2 weeks, I'm still pretty new to this.

It's about some measurements of a Diode which I get from a colleague of mine.

 

Every unit (unit no.) is measured at different temperatures. Every measurement of a unit with a different temperature should represent a line on the graph. The lines should have different colors based on the temperature but should all appear on one graph. In the end there should be one line for every unit no. measured at every temperature.

In my case, it is a VDRM vs IDRM graph. 

I will attach examples and my JMP table below for anyone interested.

"original_datatable" is the raw data that I used

"table_for_graph" is the changed table that I used to draw the graph

 

My problem here is that whenever I try to use Unit No. as Overlay and TEMP as Color, my whole graph just doesn't work anymore (see settings). I can do the different TEMPS seperated (see Graphs) but not in a single graph.

 

If any of you could help me get the correct graph in JMP I'd very much appreciate the help!

 

Kind regards!

 

1 ACCEPTED SOLUTION

Accepted Solutions

Re: Problem with graph

If I understand everything correctly, you need to concatenate the unit number with the temperature. You can only have one column as the overlay, but you want two: both the unit number and the temperature. So I created a column that converts unit number and temperature to character and joins them. Use that as your overlay column. You end up with this picture:

 

Dan_Obermiller_1-1659446642755.png

 

I have included your table with the added column and a script for the graph. I think this is what you are looking for.

Dan Obermiller

View solution in original post

3 REPLIES 3
jthi
Super User

Re: Problem with graph

One option would be to combine Unit No and TEMP into new character column which is combination of both of them

jthi_0-1659445443367.png

 

-Jarmo

Re: Problem with graph

If I understand everything correctly, you need to concatenate the unit number with the temperature. You can only have one column as the overlay, but you want two: both the unit number and the temperature. So I created a column that converts unit number and temperature to character and joins them. Use that as your overlay column. You end up with this picture:

 

Dan_Obermiller_1-1659446642755.png

 

I have included your table with the added column and a script for the graph. I think this is what you are looking for.

Dan Obermiller
Georg
Level VII

Re: Problem with graph

Adding to @jthi 's and @Dan_Obermiller 's replies, you can do exactly the same graph, with even better legend in JMP.

I simplified the legend and left Unit No apart, because due to the same color you cannot identify this in the graph.

In GraphBuilder, you can use the options: Row order( 1 ), Missing Values( "No Connection" ) to generate a line plot.

 

But the tricky part is that JMP draws a line from max(IDRM) to min(IDRM) when the next line occurrs.

That is why we need to introduce a row with missing values into the table, and sort it afterwards, to get a clean plot.

The following script shows the steps needed for this, the very low script is an independent one that contains the table for quick reproduction. Another tricky point was that there are double measurements at 25 °C, that is why Time is needed for grouping additionally.

On the other hand, this may not be the best way to visualize the data in JMP, there are a lot other representations that are meaningful and easier to do. But you asked for exactly this one.

 

// Preparing Data table for plotting lines
// 1st: generate grouping cols to separate each line
dt_sum = dt << Summary( Group( :Unit No, :TEMP, :Time ), Freq( "None" ), Weight( "None" ) );

// 2nd: Put these into the data table (with missing values for IDRM etc.)
dt << concatenate( dt_sum, append to first table );
Close( dt_sum, NoSave );

// 3rd: sort table to separate each group plotted as line by a row with missing data
dt << Sort( By( :TEMP, :Unit No, :Time, :IDRM ), Order( Ascending, Ascending, Ascending ), replace table );

// 4th: Graph Builder Script with lines for each group
Graph Builder(
	Size( 1058, 589 ),
	Show Control Panel( 0 ),
	Categorical Color Theme( "JMP Default" ),
	Graph Spacing( 5 ),
	Variables( X( :IDRM ), Y( :VDRM ), Overlay( :TEMP ) ),
	Elements( Line( X, Y, Legend( 14 ), Row order( 1 ), Missing Values( "No Connection" ) ) )
);

 

Georg_0-1659523277171.png

 

Full script:

Names Default To Here( 1 );

dt = New Table( "table_for_graph",
	Add Rows( 960 ),
	Compress File When Saved( 1 ),
	New Column( "Unit No",
		Numeric,
		"Nominal",
		Format( "Best", 12 ),
		Set Values(
			[1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1,
			1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1,
			1, 1, 1, 1, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2,
			2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2, 2,
			2, 2, 2, 2, 2, 2, 2, 2, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3,
			3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3,
			3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 3, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4,
			4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4,
			4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 4, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5,
			5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5,
			5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 5, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6,
			6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6,
			6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 6, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7,
			7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7,
			7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 7, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8,
			8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8,
			8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 8, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9,
			9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9,
			9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 9, 10, 10, 10, 10, 10, 10, 10,
			10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10,
			10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10,
			10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10, 10]
		),
		Set Display Width( 44 )
	),
	New Column( "Date",
		Numeric,
		"Continuous",
		Format( "ddMonyyyy", 10 ),
		Input Format( "ddMonyyyy" ),
		Set Values(
			[3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200, 3737059200,
			3737059200, 3737059200, 3737059200]
		),
		Set Display Width( 56 )
	),
	New Column( "Time",
		Numeric,
		"Continuous",
		Format( "h:m:s", 11, 0 ),
		Input Format( "h:m:s", 0 ),
		Set Values(
			[38012, 38012, 38012, 38012, 38012, 38012, 38012, 38012, 38012, 38012, 38012, 38012, 38012, 38012, 38012, 38012, 38184, 38184, 38184,
			38184, 38184, 38184, 38184, 38184, 38184, 38184, 38184, 38184, 38184, 38184, 38184, 38184, 38364, 38364, 38364, 38364, 38364, 38364,
			38364, 38364, 38364, 38364, 38364, 38364, 38364, 38364, 38364, 38364, 38515, 38515, 38515, 38515, 38515, 38515, 38515, 38515, 38515,
			38515, 38515, 38515, 38515, 38515, 38515, 38515, 38658, 38658, 38658, 38658, 38658, 38658, 38658, 38658, 38658, 38658, 38658, 38658,
			38658, 38658, 38658, 38658, 38863, 38863, 38863, 38863, 38863, 38863, 38863, 38863, 38863, 38863, 38863, 38863, 38863, 38863, 38863,
			38863, 39089, 39089, 39089, 39089, 39089, 39089, 39089, 39089, 39089, 39089, 39089, 39089, 39089, 39089, 39089, 39089, 39260, 39260,
			39260, 39260, 39260, 39260, 39260, 39260, 39260, 39260, 39260, 39260, 39260, 39260, 39260, 39260, 39446, 39446, 39446, 39446, 39446,
			39446, 39446, 39446, 39446, 39446, 39446, 39446, 39446, 39446, 39446, 39446, 39598, 39598, 39598, 39598, 39598, 39598, 39598, 39598,
			39598, 39598, 39598, 39598, 39598, 39598, 39598, 39598, 39742, 39742, 39742, 39742, 39742, 39742, 39742, 39742, 39742, 39742, 39742,
			39742, 39742, 39742, 39742, 39742, 39951, 39951, 39951, 39951, 39951, 39951, 39951, 39951, 39951, 39951, 39951, 39951, 39951, 39951,
			39951, 39951, 40145, 40145, 40145, 40145, 40145, 40145, 40145, 40145, 40145, 40145, 40145, 40145, 40145, 40145, 40145, 40145, 40313,
			40313, 40313, 40313, 40313, 40313, 40313, 40313, 40313, 40313, 40313, 40313, 40313, 40313, 40313, 40313, 40501, 40501, 40501, 40501,
			40501, 40501, 40501, 40501, 40501, 40501, 40501, 40501, 40501, 40501, 40501, 40501, 40655, 40655, 40655, 40655, 40655, 40655, 40655,
			40655, 40655, 40655, 40655, 40655, 40655, 40655, 40655, 40655, 40797, 40797, 40797, 40797, 40797, 40797, 40797, 40797, 40797, 40797,
			40797, 40797, 40797, 40797, 40797, 40797, 41003, 41003, 41003, 41003, 41003, 41003, 41003, 41003, 41003, 41003, 41003, 41003, 41003,
			41003, 41003, 41003, 41257, 41257, 41257, 41257, 41257, 41257, 41257, 41257, 41257, 41257, 41257, 41257, 41257, 41257, 41257, 41257,
			41428, 41428, 41428, 41428, 41428, 41428, 41428, 41428, 41428, 41428, 41428, 41428, 41428, 41428, 41428, 41428, 41615, 41615, 41615,
			41615, 41615, 41615, 41615, 41615, 41615, 41615, 41615, 41615, 41615, 41615, 41615, 41615, 41768, 41768, 41768, 41768, 41768, 41768,
			41768, 41768, 41768, 41768, 41768, 41768, 41768, 41768, 41768, 41768, 41910, 41910, 41910, 41910, 41910, 41910, 41910, 41910, 41910,
			41910, 41910, 41910, 41910, 41910, 41910, 41910, 42118, 42118, 42118, 42118, 42118, 42118, 42118, 42118, 42118, 42118, 42118, 42118,
			42118, 42118, 42118, 42118, 42334, 42334, 42334, 42334, 42334, 42334, 42334, 42334, 42334, 42334, 42334, 42334, 42334, 42334, 42334,
			42334, 42506, 42506, 42506, 42506, 42506, 42506, 42506, 42506, 42506, 42506, 42506, 42506, 42506, 42506, 42506, 42506, 42688, 42688,
			42688, 42688, 42688, 42688, 42688, 42688, 42688, 42688, 42688, 42688, 42688, 42688, 42688, 42688, 42839, 42839, 42839, 42839, 42839,
			42839, 42839, 42839, 42839, 42839, 42839, 42839, 42839, 42839, 42839, 42839, 42982, 42982, 42982, 42982, 42982, 42982, 42982, 42982,
			42982, 42982, 42982, 42982, 42982, 42982, 42982, 42982, 43188, 43188, 43188, 43188, 43188, 43188, 43188, 43188, 43188, 43188, 43188,
			43188, 43188, 43188, 43188, 43188, 44196, 44196, 44196, 44196, 44196, 44196, 44196, 44196, 44196, 44196, 44196, 44196, 44196, 44196,
			44196, 44196, 44367, 44367, 44367, 44367, 44367, 44367, 44367, 44367, 44367, 44367, 44367, 44367, 44367, 44367, 44367, 44367, 44549,
			44549, 44549, 44549, 44549, 44549, 44549, 44549, 44549, 44549, 44549, 44549, 44549, 44549, 44549, 44549, 44701, 44701, 44701, 44701,
			44701, 44701, 44701, 44701, 44701, 44701, 44701, 44701, 44701, 44701, 44701, 44701, 44842, 44842, 44842, 44842, 44842, 44842, 44842,
			44842, 44842, 44842, 44842, 44842, 44842, 44842, 44842, 44842, 45049, 45049, 45049, 45049, 45049, 45049, 45049, 45049, 45049, 45049,
			45049, 45049, 45049, 45049, 45049, 45049, 45261, 45261, 45261, 45261, 45261, 45261, 45261, 45261, 45261, 45261, 45261, 45261, 45261,
			45261, 45261, 45261, 45431, 45431, 45431, 45431, 45431, 45431, 45431, 45431, 45431, 45431, 45431, 45431, 45431, 45431, 45431, 45431,
			45620, 45620, 45620, 45620, 45620, 45620, 45620, 45620, 45620, 45620, 45620, 45620, 45620, 45620, 45620, 45620, 45773, 45773, 45773,
			45773, 45773, 45773, 45773, 45773, 45773, 45773, 45773, 45773, 45773, 45773, 45773, 45773, 45916, 45916, 45916, 45916, 45916, 45916,
			45916, 45916, 45916, 45916, 45916, 45916, 45916, 45916, 45916, 45916, 46121, 46121, 46121, 46121, 46121, 46121, 46121, 46121, 46121,
			46121, 46121, 46121, 46121, 46121, 46121, 46121, 46335, 46335, 46335, 46335, 46335, 46335, 46335, 46335, 46335, 46335, 46335, 46335,
			46335, 46335, 46335, 46335, 46507, 46507, 46507, 46507, 46507, 46507, 46507, 46507, 46507, 46507, 46507, 46507, 46507, 46507, 46507,
			46507, 46685, 46685, 46685, 46685, 46685, 46685, 46685, 46685, 46685, 46685, 46685, 46685, 46685, 46685, 46685, 46685, 46837, 46837,
			46837, 46837, 46837, 46837, 46837, 46837, 46837, 46837, 46837, 46837, 46837, 46837, 46837, 46837, 46980, 46980, 46980, 46980, 46980,
			46980, 46980, 46980, 46980, 46980, 46980, 46980, 46980, 46980, 46980, 46980, 47188, 47188, 47188, 47188, 47188, 47188, 47188, 47188,
			47188, 47188, 47188, 47188, 47188, 47188, 47188, 47188, 47393, 47393, 47393, 47393, 47393, 47393, 47393, 47393, 47393, 47393, 47393,
			47393, 47393, 47393, 47393, 47393, 47563, 47563, 47563, 47563, 47563, 47563, 47563, 47563, 47563, 47563, 47563, 47563, 47563, 47563,
			47563, 47563, 47750, 47750, 47750, 47750, 47750, 47750, 47750, 47750, 47750, 47750, 47750, 47750, 47750, 47750, 47750, 47750, 47901,
			47901, 47901, 47901, 47901, 47901, 47901, 47901, 47901, 47901, 47901, 47901, 47901, 47901, 47901, 47901, 48043, 48043, 48043, 48043,
			48043, 48043, 48043, 48043, 48043, 48043, 48043, 48043, 48043, 48043, 48043, 48043, 48249, 48249, 48249, 48249, 48249, 48249, 48249,
			48249, 48249, 48249, 48249, 48249, 48249, 48249, 48249, 48249, 48557, 48557, 48557, 48557, 48557, 48557, 48557, 48557, 48557, 48557,
			48557, 48557, 48557, 48557, 48557, 48557, 48731, 48731, 48731, 48731, 48731, 48731, 48731, 48731, 48731, 48731, 48731, 48731, 48731,
			48731, 48731, 48731, 48917, 48917, 48917, 48917, 48917, 48917, 48917, 48917, 48917, 48917, 48917, 48917, 48917, 48917, 48917, 48917,
			49067, 49067, 49067, 49067, 49067, 49067, 49067, 49067, 49067, 49067, 49067, 49067, 49067, 49067, 49067, 49067, 49209, 49209, 49209,
			49209, 49209, 49209, 49209, 49209, 49209, 49209, 49209, 49209, 49209, 49209, 49209, 49209, 49414, 49414, 49414, 49414, 49414, 49414,
			49414, 49414, 49414, 49414, 49414, 49414, 49414, 49414, 49414, 49414]
		),
		Set Display Width( 116 )
	),
	New Column( "TEMP",
		Numeric,
		"Continuous",
		Format( "Best", 12 ),
		Set Values(
			[25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40,
			-40, -40, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125,
			125, 125, 125, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25,
			25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40,
			-40, -40, -40, -40, -40, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125,
			125, 125, 125, 125, 125, 125, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 25, 25, 25, 25, 25, 25, 25,
			25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, -40, -40, -40, -40, -40, -40, -40,
			-40, -40, -40, -40, -40, -40, -40, -40, -40, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 125, 125, 125, 125, 125, 125,
			125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 25, 25,
			25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, -40, -40, -40,
			-40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 125, 125,
			125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150,
			150, 150, 150, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25,
			25, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85,
			85, 85, 85, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 150, 150, 150, 150, 150, 150, 150, 150, 150,
			150, 150, 150, 150, 150, 150, 150, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25,
			25, 25, 25, 25, 25, 25, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, 85, 85, 85, 85, 85, 85, 85, 85,
			85, 85, 85, 85, 85, 85, 85, 85, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 150, 150, 150, 150, 150,
			150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25,
			25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, 85, 85, 85,
			85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 150,
			150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25,
			25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40,
			-40, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125,
			125, 125, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25,
			25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40, -40,
			-40, -40, -40, -40, -40, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 125, 125, 125, 125, 125, 125, 125, 125, 125, 125,
			125, 125, 125, 125, 125, 125, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 25, 25, 25, 25, 25, 25, 25,
			25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, -40, -40, -40, -40, -40, -40, -40,
			-40, -40, -40, -40, -40, -40, -40, -40, -40, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 85, 125, 125, 125, 125, 125, 125,
			125, 125, 125, 125, 125, 125, 125, 125, 125, 125, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 150, 25, 25,
			25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25, 25]
		),
		Set Display Width( 38 )
	),
	New Column( "VDRM",
		Numeric,
		"Continuous",
		Format( "Best", 12 ),
		Set Values(
			[1360.823, 1447.839, 1468.9561, 1136.908, 1362.66, 1225.1479, 1371.637, 1288.599, 1378.165, 1390.611, 1365.108, 1409.177, 1409.993,
			1431.824, 1415.4, 1421.827, 1350.418, 1227.29, 1215.967, 943.49323, 1152.413, 1040.302, 1204.1331, 1213.5179, 1171.286, 1178.426,
			1193.626, 1257.485, 1246.5699, 1252.181, 1254.425, 1258.302, 0.013536, 0.19048689, 0.59853399, 3.556875, 20.490829, 77.107353, 300.00311,
			619.09583, 851.98871, 1048.769, 1206.582, 1329.4041, 1392.345, 1458.8571, 1495.5811, 1519.452, 0.088475, 0.088475, 0.088475, 0.088475,
			0.088475, 0.088475, 0.013536, 0.29249859, 0.39451051, 2.0266991, 17.02243, 81.289841, 155.45239, 297.6568, 433.43439, 553.60431, 0.013536,
			0.088475, 0.088475, 0.19048689, 0.088475, 0.19048689, 0.088475, 0.088475, 0.088475, 0.19048689, 0.088475, 0.088475, 0.19048689,
			0.39451051, 2.7407811, 12.12586, 1375.819, 1457.939, 1487.42, 1489.868, 1490.786, 1491.807, 1321.7531, 1348.072, 1329.098, 1376.0229,
			1399.486, 1429.783, 1419.99, 1419.786, 1427.0291, 1431.416, 1391.223, 1477.1169, 1515.0649, 1520.9821, 1523.328, 1524.552, 1428.559,
			1432.844, 1333.688, 1356.437, 1419.072, 1472.22, 1451.41, 1462.937, 1464.671, 1469.976, 1434.6801, 1412.951, 1383.674, 1379.39, 1350.928,
			1261.26, 1098.653, 1250.957, 1201.481, 1185.363, 1213.62, 1247.488, 1250.957, 1249.019, 1249.733, 1251.7729, 0.088475, 0.19048689,
			1.312616, 7.6373448, 35.384541, 111.2813, 381.0004, 723.35193, 952.57233, 1132.623, 1272.073, 1380.2061, 1436.21, 1496.8051, 1530.469,
			1552.912, 0.088475, 0.088475, 0.19048689, 0.088475, 0.013536, 0.013536, 0.19048689, 0.19048689, 0.49652219, 3.556875, 25.591419, 104.6505,
			190.9525, 351.31491, 495.96771, 620.62598, 0.013536, 0.088475, 0.088475, 0.013536, 0.19048689, 0.088475, 0.088475, 0.013536, 0.013536,
			0.088475, 0.088475, 0.19048689, 0.19048689, 0.49652219, 4.0669341, 18.348579, 1402.954, 1483.952, 1522.2061, 1531.693, 1534.55, 1536.08,
			1538.0179, 1412.339, 1356.233, 1381.4301, 1389.489, 1464.569, 1465.896, 1473.853, 1471.098, 1475.281, 1381.634, 1469.058, 1509.863,
			1517.0031, 1518.738, 1519.452, 1282.886, 1372.453, 1393.671, 1411.319, 1395.507, 1462.121, 1461.713, 1470.486, 1476.6071, 1474.465,
			1431.416, 1409.891, 1395.099, 1385.408, 1090.288, 1016.636, 1228.922, 1162.5129, 1222.394, 1200.9709, 1244.02, 1273.705, 1274.215,
			1274.011, 1276.562, 1280.642, 0.19048689, 0.19048689, 0.80255753, 5.3930869, 29.05982, 95.367462, 342.64401, 672.03992, 901.0564,
			1087.942, 1235.859, 1351.54, 1411.421, 1475.791, 1511.903, 1535.672, 0.088475, 0.19048689, 0.088475, 0.088475, 0.088475, 0.013536,
			0.088475, 0.088475, 0.59853399, 2.6387689, 20.490829, 89.144737, 164.9395, 313.87659, 452.9187, 573.70062, 0.19048689, 0.013536, 0.013536,
			0.013536, 0.088475, 0.013536, 0.088475, 0.2175601, 0.013536, 0.013536, 0.088475, 0.19048689, 0.088475, 0.59853399, 2.7407811, 13.35,
			1388.469, 1473.546, 1515.575, 1528.021, 1531.183, 1532.8149, 1409.483, 1496.0909, 1358.375, 1392.651, 1402.1379, 1475.791, 1476.913,
			1475.995, 1482.524, 1482.524, 1387.856, 1473.75, 1500.988, 1503.538, 1504.762, 1505.476, 1285.947, 1340.7271, 1330.832, 1357.457,
			1426.825, 1456.6121, 1445.901, 1451.104, 1455.184, 1461.101, 1414.788, 1400.71, 1389.183, 1373.0649, 1163.941, 1145.783, 1187.811,
			1224.1281, 1174.958, 1221.271, 1226.27, 1258.4041, 1260.8521, 1258.506, 1260.444, 1262.892, 0.088475, 0.39451051, 1.210605, 7.3313098,
			33.956379, 108.425, 371.51331, 707.23401, 935.74042, 1118.137, 1260.75, 1371.943, 1429.4771, 1491.603, 1525.879, 1548.627, 0.088475,
			0.088475, 0.088475, 0.088475, 0.1155484, 0.088475, 0.013536, 0.19048689, 0.49652219, 2.9448049, 22.225031, 96.183563, 180.2412, 332.64679,
			472.19891, 594.20502, 0.088475, 0.088475, 0.013536, 0.19048689, 0.19048689, 0.088475, 0.013536, 0.088475, 0.013536, 0.088475, 0.088475,
			0.19048689, 0.088475, 0.59853399, 3.2508399, 15.08421, 1400.812, 1481.809, 1516.085, 1520.574, 1522.002, 1522.818, 1278.3979, 1324.609,
			1430.906, 1442.127, 1388.877, 1474.771, 1459.775, 1467.0179, 1465.7939, 1469.67, 1378.267, 1466.304, 1497.519, 1482.7271, 1301.657,
			1197.401, 1294.822, 1398.772, 1417.542, 1381.226, 1392.651, 1427.743, 1429.4771, 1434.6801, 1441.515, 1446.207, 1408.871, 1251.161,
			1148.129, 1299.514, 996.64142, 1269.625, 1097.123, 1203.011, 1232.697, 1236.675, 1230.7581, 1227.494, 1226.474, 1229.126, 1234.431,
			1239.837, 0.088475, 0.29249859, 1.0065809, 5.8011341, 28.65177, 96.183563, 340.70569, 652.35162, 878.61377, 1067.3361, 1219.741, 1339.299,
			1401.118, 1467.222, 1503.64, 1527.4091, 0.088475, 0.013536, 0.088475, 0.088475, 0.088475, 0.013536, 0.19048689, 0.19048689, 0.19048689,
			2.12871, 17.328461, 82.513977, 161.9812, 306.32779, 433.8425, 549.82977, 0.088475, 0.1155484, 0.013536, 0.088475, 0.013536, 0.088475,
			0.013536, 0.088475, 0.088475, 0.088475, 0.088475, 0.088475, 0.19048689, 0.39451051, 2.230722, 10.6977, 1390.917, 1471.506, 1505.782,
			1514.249, 1517.718, 1519.7581, 1520.676, 1315.8361, 1460.795, 1393.671, 1438.964, 1472.322, 1461.611, 1463.651, 1462.835, 1464.7729,
			1326.854, 1389.285, 1395.201, 1396.936, 1398.4659, 1399.282, 1143.2321, 1194.442, 1244.53, 1287.069, 1311.45, 1348.072, 1345.42, 1338.585,
			1349.6021, 1352.255, 1319.203, 1289.2111, 1268.299, 1255.241, 1242.49, 1201.277, 1147.007, 1110.385, 1117.933, 1118.24, 1116.403,
			1125.584, 1122.728, 1124.666, 1131.501, 1137.52, 0.013536, 0.19048689, 1.0065809, 6.4132051, 30.385969, 94.245331, 385.28491, 751.20099,
			983.88989, 1149.863, 1267.891, 1354.8051, 1399.486, 1446.921, 1472.322, 1482.217, 0.013536, 0.013536, 0.013536, 0.088475, 0.088475,
			0.19048689, 0.088475, 0.19048689, 0.49652219, 3.454864, 25.38739, 101.7942, 194.8289, 371.92139, 527.79529, 662.34882, 0.19048689,
			0.088475, 0.19048689, 0.013536, 0.013536, 0.088475, 0.19048689, 0.013536, 0.088475, 0.088475, 0.088475, 0.19048689, 0.088475, 0.80255753,
			4.3729701, 18.858641, 1347.562, 1407.3409, 1412.7469, 1414.38, 1415.196, 1416.012, 1153.842, 1213.314, 1260.036, 1299.718, 1296.046,
			1345.522, 1340.5229, 1348.99, 1356.131, 1358.783, 1374.493, 1453.348, 1490.276, 1498.947, 1500.478, 1147.415, 1356.947, 1315.224,
			1420.092, 1432.028, 1413.666, 1431.62, 1430.498, 1442.229, 1441.719, 1454.062, 1429.069, 1414.278, 1387.755, 1377.349, 1366.842, 1346.746,
			1113.751, 1114.261, 1129.359, 1182.303, 1179.344, 1216.375, 1215.661, 1217.4969, 1217.191, 1219.741, 0.19048689, 0.19048689, 1.618652,
			8.8614864, 38.240879, 108.629, 433.3324, 815.46838, 1047.239, 1208.316, 1322.365, 1407.647, 1451.512, 1499.764, 1526.899, 1544.955,
			0.013536, 0.088475, 0.19048689, 0.088475, 0.19048689, 0.013536, 0.19048689, 0.19048689, 0.80255753, 5.3930869, 33.548328, 120.5644,
			222.7802, 420.07089, 586.34998, 726.10608, 0.013536, 0.19048689, 0.19048689, 0.088475, 0.088475, 0.088475, 0.013536, 0.19048689, 0.088475,
			0.088475, 0.088475, 0.19048689, 0.19048689, 0.90456933, 7.2292991, 27.73366, 1385.6121, 1460.5909, 1495.989, 1503.7419, 1505.068, 1505.17,
			1363.578, 1307.777, 1396.7321, 1423.459, 1396.63, 1439.679, 1445.595, 1440.495, 1453.6541, 1454.47, 1377.859, 1463.141, 1497.723,
			1502.926, 1504.864, 1505.986, 1269.115, 1329.608, 1334.198, 1358.171, 1409.483, 1445.799, 1441.413, 1449.9821, 1454.368, 1462.223,
			1428.7629, 1406.015, 1390.713, 1074.17, 1350.826, 1349.908, 1094.879, 1254.833, 1155.168, 1219.4351, 1185.873, 1243.918, 1241.062,
			1241.1639, 1241.47, 1244.02, 0.088475, 0.19048689, 0.90456933, 4.883028, 24.16325, 82.513977, 339.99161, 686.52557, 924.11102, 1113.241,
			1258.506, 1369.392, 1425.601, 1485.686, 1519.248, 1541.1801, 0.088475, 0.088475, 0.088475, 0.013536, 0.013536, 0.088475, 0.19048689,
			0.088475, 0.49652219, 2.842793, 22.021, 92.817169, 172.9984, 344.9902, 496.78381, 625.93073, 0.013536, 0.19048689, 0.013536, 0.088475,
			0.013536, 0.088475, 0.19048689, 0.088475, 0.088475, 0.088475, 0.088475, 0.19048689, 0.19048689, 0.59853399, 3.556875, 16.41036, 1391.121,
			1471.302, 1508.333, 1516.799, 1518.228, 1518.636, 1405.097, 1373.575, 1464.161, 1412.951, 1436.72, 1458.347, 1460.0811, 1463.243,
			1466.916, 1470.1801, 1321.141, 1393.059, 1396.7321, 1183.9351, 1365.618, 1205.358, 1352.356, 1218.415, 1250.0389, 1276.562, 1339.197,
			1361.333, 1372.453, 1367.25, 1373.983, 1378.165, 1340.319, 1320.121, 848.62231, 945.02338, 1067.03, 926.55933, 1155.3719, 1179.0389,
			1092.941, 1119.974, 1147.517, 1184.751, 1185.057, 1187.403, 1189.954, 1196.074, 0.088475, 0.19048689, 1.108593, 4.576993, 23.143129,
			76.393272, 330.19849, 674.28418, 911.86963, 1095.593, 1231.166, 1331.444, 1382.45, 1436.516, 1466.61, 1486.298, 0.088475, 0.013536,
			0.19048689, 0.088475, 0.088475, 0.013536, 0.088475, 0.19048689, 0.29249859, 2.230722, 16.716391, 76.495293, 153.3102, 308.9801, 451.49051,
			577.67902, 0.013536, 0.013536, 0.088475, 0.013536, 0.013536, 0.088475, 0.013536, 0.19048689, 0.088475, 0.19048689, 0.088475, 0.088475,
			0.19048689, 0.59853399, 2.434746, 11.30977, 1334.8101, 1412.543, 1442.3311, 1445.085, 1446.207, 1446.615, 1212.396, 1396.834, 1284.723,
			1322.7729, 1334.402, 1386.53, 1383.572, 1393.569, 1394.283, 1400.098, 1333.892, 1383.572, 1392.549, 1396.12, 1398.568, 1401.016, 1276.358,
			1350.1121, 1315.224, 1285.0291, 1286.049, 1348.582, 1352.458, 1356.845, 1359.803, 1367.454, 1085.5959, 1333.076, 1313.49, 1288.4969,
			1276.97, 1266.769, 1027.959, 1064.887, 1130.277, 1109.875, 1111.7111, 1157.8199, 1157.106, 1155.678, 1157.004, 1159.554, 0.19048689,
			0.19048689, 1.414628, 8.045393, 36.81271, 111.3833, 424.86539, 799.55463, 1023.776, 1178.63, 1287.681, 1368.984, 1410.5031, 1454.776,
			1477.933, 1491.5, 0.088475, 0.088475, 0.088475, 0.19048689, 0.088475, 0.088475, 0.19048689, 0.19048689, 0.59853399, 4.576993, 32.120171,
			120.2583, 222.8822, 408.74759, 568.59998, 704.2757, 0.013536, 0.013536, 0.19048689, 0.088475, 0.088475, 0.1155484, 0.013536, 0.088475,
			0.19048689, 0.19048689, 0.19048689, 0.088475, 0.29249859, 0.70054579, 5.5971112, 23.653191, 1348.48, 1410.809, 1425.601, 1429.171,
			1431.313, 1433.15, 1180.569, 1339.197, 1310.022, 1273.399, 1335.3199, 1360.7209, 1365.006, 1375.5129, 1375.615, 1381.4301]
		),
		Set Display Width( 71 )
	),
	New Column( "IDRM",
		Numeric,
		"Continuous",
		Format( "Engineering", 12 ),
		Set Values(
			[0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004,
			0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003,
			0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002,
			0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015,
			0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001,
			0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007,
			0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005,
			0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035,
			0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002,
			0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001,
			0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007,
			0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005,
			0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035,
			0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002,
			0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001,
			0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005,
			0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004,
			0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003,
			0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002,
			0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015,
			0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001,
			0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007,
			0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005,
			0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035,
			0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002,
			0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001,
			0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007,
			0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005,
			0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035,
			0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002,
			0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001,
			0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005,
			0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004,
			0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003,
			0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002,
			0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015,
			0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001,
			0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007,
			0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005,
			0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035,
			0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002,
			0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001,
			0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007,
			0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005,
			0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035,
			0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002,
			0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001,
			0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005,
			0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004,
			0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003,
			0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002,
			0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015,
			0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001,
			0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007,
			0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005,
			0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035,
			0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002,
			0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007, 0.00001,
			0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005, 0.000007,
			0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035, 0.000005,
			0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002, 0.0000035,
			0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001, 0.000002,
			0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005, 0.000001,
			0.000002, 0.0000035, 0.000005, 0.000007, 0.00001, 0.00002, 0.000035, 0.00005, 0.00007, 0.0001, 0.00015, 0.0002, 0.0003, 0.0004, 0.0005]
		),
		Set Display Width( 46 )
	),
	Set Row States(
		[0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
		0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 0, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1,
		1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1,
		1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1, 1]
	)
);

// Preparing Data table for plotting lines
// 1st: generate grouping cols to separate each line
dt_sum = dt << Summary( Group( :Unit No, :TEMP, :Time ), Freq( "None" ), Weight( "None" ) );

// 2nd: Put these into the data table (with missing values for IDRM etc.)
dt << concatenate( dt_sum, append to first table );
Close( dt_sum, NoSave );

// 3rd: sort table to separate each group plotted as line by a row with missing data
dt << Sort( By( :TEMP, :Unit No, :Time, :IDRM ), Order( Ascending, Ascending, Ascending ), replace table );

// 4th: Graph Builder Script with lines for each group
Graph Builder(
	Size( 1058, 589 ),
	Show Control Panel( 0 ),
	Categorical Color Theme( "JMP Default" ),
	Graph Spacing( 5 ),
	Variables( X( :IDRM ), Y( :VDRM ), Overlay( :TEMP ) ),
	Elements( Line( X, Y, Legend( 14 ), Row order( 1 ), Missing Values( "No Connection" ) ) )
);
Georg

Recommended Articles