jjhartwi
Gauge variation, linearity and bias
When you perform a gauge linearity study and the "mean" bias increases as you inrease along the range of measurements you are said to have a "linearity" problem with the gauge as indicated by the positive slope in the bias-scale graph.  If the bias graph slope is zero, random bias, then you are said not to have a linearity problem with the gauge.  But what do you can a graph with no linearity problem of the "average" bias over the range of the gauge but you do have increasing "variation" of the multiple measurements you take along the scale?  In other words, if you take 10 different measurements at each point along the scale, the graph looks like the measurements are included in a in a cone shape.  Do you just call it a gauge repeatability problem with a special cause of variation or is there a name for it?
Increasing VariationIncreasing Variation  Non-biasNon-bias
Re: Gauge variation, linearity and bias
It's a good question! You are correct that technically, in the measurement standard documents, linearity is defined as the absence of variability due to bias over the measurement range. In the JMP course notes for the MSA course, it says "Some authors include the absence of variability due to reproducibility over the measurement range." In my experience, companies that notice this problem call it linearity, and other companies don't notice the problem, so don't even ask the question. I have not heard of another term for it, in MSA particularly.
jjhartwi
Here is my problem.  A gauge R&R is based on ANOVA, either 2 way or 3 way.  One of the ANOVA's assumption is equal variances.  So in the case where the variance changes over the range of the measurement device, will this cause a problem with the gauge R&R results?
Short answer: no.
Long answer: The ANOVA you are fitting is probably a random effects ANOVA, with the Part effect being random. Maybe you have other random terms as well, such as time effects, operator effects, materials, gauges, setups, etc. Maybe one of those effects should be the nominal value of the part. Then you can model the variance as a function of the nominal value (and other terms) and you don't have to assume equal variances. If the nominal values are really far apart, consider doing a separate analysis for each range. For example, if you have parts at 10, 1000, and 10000, you might perform three separate analyses.
In JMP, if all your effects are random, you can use the EMP or Variability platforms to get those variance components. If some effects are fixed and others are random, use Analyze > Fit Model to specify the mixed model and get the variance components.
I hope this helps!