cancel
Showing results for 
Show  only  | Search instead for 
Did you mean: 
  • Sign-in to the JMP Community will be unavailable intermittently Dec. 6-7 due to a system update. Thank you for your understanding!
  • We’re retiring the File Exchange at the end of this year. The JMP Marketplace is now your destination for add-ins and extensions.
  • JMP 19 is here! Learn more about the new features.

Discussions

Solve problems, and share tips and tricks with other JMP users.
Choose Language Hide Translation Bar
Rily_Maya
Level III

Degradation Data Analysis, Ea<0, is there a problem?

1.png..JMP\JMPPRO\18\Samples\Data\Reliability\Device B.jmp, Degradation Data Analysis, Ea<0, is there a problem?

7 REPLIES 7
jthi
Super User

Re: Degradation Data Analysis, Ea<0, is there a problem?

Rily_Maya
Level III

Re: Degradation Data Analysis, Ea<0, is there a problem?

I have read it, but I still don't understand the above questions.

Re: Degradation Data Analysis, Ea<0, is there a problem?

Hi @Rily_Maya ,

 

In the approach you're doing, it isn't appropriate to try to fit the model per system ID - there is not sufficient information or variation around the parameters to correctly define the activation energy (Ea) for each system, because each system hasn't had a number of temperatures tested against them - instead the intent is to apply the model to the whole data set to find a suitably fitting model to the data (see below) that can generalise against the variation. The data isn't perfect (I suspect intentionally), so in real world settings you would maybe consider applying a different formula, or refining the data that is available. Another option would be to take the formula into the Non-linear platform to add more restrictions and perform things like SSE grid searches to better refine the parameters. 

 

Ben_BarrIngh_0-1754463643238.png

Ben_BarrIngh_1-1754463664462.png

 

Thanks,
Ben

 

 

“All models are wrong, but some are useful”
Rily_Maya
Level III

Re: Degradation Data Analysis, Ea<0, is there a problem?

If using Device B.jmp as an example in "Reliability and Survival Methods > Degradation > Model Specification > Nonlinear Path" is misleading, then the corresponding example is inappropriate.

Re: Degradation Data Analysis, Ea<0, is there a problem?

Hi @Rily_Maya,

 

The Device B set is intended to show a modelling approach that separates the reaction rate model per temperature, allowing for estimation of an average parameter estimate for the Ea and other parameters - in this context the goal isn't to look at each device as they aren't tested that way - the approach is appropriate and includes noise and other variations that you would expect in the real world.  What's the goal for going through the analysis? For education or for practice?

 

Thanks,

Ben

“All models are wrong, but some are useful”
Rily_Maya
Level III

Re: Degradation Data Analysis, Ea<0, is there a problem?

Set the lower specification limit to -0.5 to calculate the pseudo-failure time of each path after 130,000 hours under the operating condition of 80°C, and this is also for learning and research purposes.

Re: Degradation Data Analysis, Ea<0, is there a problem?

Hi @Rily_Maya ,

 

In that case, you want to use the Destructive Degradation platform instead, so that you can relate the prediction to the temperature that is used and not the specific equipment. Open up the platform and fill it as so.

Ben_BarrIngh_0-1754477501817.png

 

Once that’s run, go to Red Button > Prediction Settings and apply your baseline to the process, then run the model as you have done.

Ben_BarrIngh_1-1754477501822.png

 

Then you will see your inverse predictions for 80C in the report tabs below, you can click the red button here and ‘save crossing time’

Ben_BarrIngh_2-1754477501827.png

 

The 'after 130,000 hours' part doesn't play into here because you're looking for an inverse prediction of the temperature, to provide a failure time.

 

Thanks,

Ben

 

“All models are wrong, but some are useful”