cancel
Showing results for 
Show  only  | Search instead for 
Did you mean: 
Try the Materials Informatics Toolkit, which is designed to easily handle SMILES data. This and other helpful add-ins are available in the JMP® Marketplace
Choose Language Hide Translation Bar
BigJasonB
Level I

Variance Components & Spec Setting

Hello all,

 

I am trying to set a bias specification for samples run on an ELISA test involving calibrations. I know I get calibration to calibration and run to run (within calibration) variation in the observed bias when I run, say, 20 samples. 

 

 I want to explore both cal to cal and run to run variation and use the info to set an appropriate spec. I am having trouble finding a resource to read up on how to properly set up and experiment and utilize the data to create a bias specification...

 

I know I want to do something like this:

10 calibrations run on one day

2 runs of 20 samples per calibration (total of 20 runs)

 

I want to then assess the variation I see and set specs around that. 

 

I know it's not specifically a JMP question but I am using JMP for all of this. Any help is appreciated or maybe just suggestions for reading? Sorry I am quite new to this all and a bit lost where to start. 

 

 

 

 

 

 

2 REPLIES 2

Re: Variance Components & Spec Setting

Bias is generally understood to mean a fixed effect on the response, like non-linearity. Your effects, such as repeated calibrations, are random effects. Sometimes in assay development there is a distinction between accuracy (lack of bias) and precision (cumulative random effects). Other situations consider accuracy to be the combination of the bias and the random effects.

 

The easiest way to design such a study with JMP is DOE > Classical > Full Factorial Design. Enter all your factors as categorical.

 

Are your 20 samples the same control, different controls, or random patient samples? That is, are you replicating the assay 20 times on the same sample or running different samples? If different samples, is there any replication?

 

The analysis will require a few extra steps. Select Calibration in the column list and Run in the Effects list and click Nest. Select Calibration and Run in the column list and Sample in the Effects list and click Nest. Select all the terms in the Effects list, click the red triangle next to Attributes, and select Random Effect. Your Fit Model dialog should look like this:

 

Screen Shot 2019-04-12 at 7.29.43 AM.png

 

I made a data table as I described above with the updated Model table script and attached it to my reply for your examination.

BigJasonB
Level I

Re: Variance Components & Spec Setting

Ok wonderful, thanks! This is a great start for me. I need to think about the reps and play around with setting up and will follow up once I do that. Might be a few days.