cancel
Showing results for 
Show  only  | Search instead for 
Did you mean: 
Submit your abstract to the call for content for Discovery Summit Americas by April 23. Selected abstracts will be presented at Discovery Summit, Oct. 21- 24.
Choose Language Hide Translation Bar
ElenaB
Level I

Interpretation of JMP GR&R result table in JMP 15.1.0

Hi Community of Experienced Users!

I am a new to JMP and  trying to interpret the Gauge R&R results.

When I look at the Result table: could you explain why Reproducibility is 60.54 %, if the formula for calculation is V(total)- V(Assay) -V(Within)? When I apply this formula, I am getting: 63.63 (Total)-3.61 (Assay) - 19.26 ( Within) =40.76.

Could you please help? The second question is what is categorized into "equipment variation"? Thank you!

ElenaB_0-1603499825712.png

 

2 REPLIES 2

Re: Interpretation of JMP GR&R result table in JMP 15.1.0

Did you check the help for Gage R&R yet?

shampton82
Level VII

Re: Interpretation of JMP GR&R result table in JMP 15.1.0

It is because the Stdev’s don’t add up to 100%. If you want to look at the % of variation from each component in a way that’s adds up to 100% use the variance component table at the bottom of the output or click on the variance output table for more detailed breakout.