Thanks for the response. I now realize I should clarify that the goal is to test the final assembled configuration of the device, since that’s what the agency will evaluate.
Regarding your question about agency expectations, Section III of the FDA’s Endotoxins Testing for Medical Devices: Questions and Answers guidance outlines some key principles. It emphasizes the importance of demonstrating that devices consistently meet endotoxin limits, and specifically calls for testing across multiple lots and component types to ensure reproducibility. It also notes that testing should be performed on the final configuration of the device as it will be used clinically, which is a critical consideration for our study design.
I'm not 100% sure what you mean by using a hypothesis, we do expect the different components to vary some which is why we need to do this study but don't have any reason to believe one component or lot would be more than another. We essentially want to prove that any combination would see have us below spec. We do plan to test 3 different lots of the 3 different types of the 3 different components but of course are not able to test every iteration of that so looking for the most appropriate way to come up with those combinations/study.